3
$\begingroup$

I am attempting to create a simple Abbe condenser for a microscope, which typically consists of two converging lenses.

Ideally, I would like to achieve a numerical aperture for the condenser of between 0.75 and 0.8, but all the single element lenses I could use to build the system have much lower NAs. I thought that the NA of a compound lens system (for imaging) would be limited to its’ lowest element’s NA, but this seems incorrect for illumination as Abbe condensers regularly are rated for NAs of >1 in oil.

In a multi-element lens system, can the NA of the system be greater than the NA of any one element in the system?

$\endgroup$
1

1 Answer 1

1
$\begingroup$

If you go to the wikipedia page about numerical aperture, you will see that the definition has nothing to do with lenses, but with the half-angle of the light cone ($\theta$, assuming everything in air):

$NA=sin(\theta)$

From here it is "easy" to understand that for a single lens one can approximate it as:

$NA\approx\frac{D}{2f}$

With $D$ the diameter of the lens and $f$ its focal length.

However, as soon as you put two lenses in a row, the definition has to be done for the light cone, rather than for a single element.

I will however follow up with the simplest case, using approximations, to show that the NA can be much higher. Imagine two lenses with the same diameter ($D_1=D_2=D$), same focal length ($f_1=f_2=f$), separated by 1/4 of their original focal length. We can use the effective focal length formula to find the new $f_n$ effective focal length of such a system:

$\frac{1}{f_n}=\frac{1}{f_1}+\frac{1}{f_2}-\frac{d}{f_1f_2}$

with $d$ as the separation between the two lenses. Then we get:

$f_n=f/1.75$

which in turn, using the approximation $NA\approx\frac{D}{2f}$ means that: $NA_n=1.75*NA_o$, where $NA_n$ is the new effective NA of your system when compared to the original of a single lens $NA_o$. Not only is the new NA much higher, the second lens can also be down to $D_2=\frac{3}{4}D$. As long as $D_2>\frac{3}{4}D$ there is no influence whatsoever on the resulting NA. Being 3/4 smaller means that its NA is also 3/4 of the original.

As an exercise, draw the system I just described and you will see that the half-angle is much steeper for the new system.

So, in conclusion, yes, the NA is defined from the half-angle and it has nothing to do with the single lens NA of a compound system, which can in the end be effectively much higher.

$\endgroup$

Your Answer

By clicking “Post Your Answer”, you agree to our terms of service and acknowledge you have read our privacy policy.

Not the answer you're looking for? Browse other questions tagged or ask your own question.