I am trying to measure the MTF of a radiology imaging system from a set of CT scans of a phantom. The imaging system is primarily designed for stationary scans so the CT images were low in spatial resolution. I am not particularly familiar with optics and I cannot come to a physical understanding of the results I produced. I am hoping that you could help me. Here is what I had done:
I have measured a line profile and calculated its modulation transfer function (MTF) according to the following: $$MTF(f)=|\int LSF(x)e^{-2\pi ifx}dx|$$
The profile was taken using Image J in a similar fashion as depicted below:
The line spread function was calculated after basic baseline removal:
Fourier transform of the profile was performed in Mathematica, MTF was then normalised to 100% and plotted:
There are several things in this plot that I do not understand.
- the first data point has value < 1
- the plot is symmetric about a point. Only half of the plot is useful?
- why does MTF reach zero twice before the point of symmetry? Is this because of the asymmetry in LSF?
- LSF(x) was a plot of intensity (grey value) varying with position (mm), would the spatial frequency in MTF(f) have units of mm^-1?
When measuring MTF using 14 line pairs (14 line profiles drawn, max&min intensity values were tabulated to produce a plot of %amplitude retained by the system), the following was produced:
and this was a much nicer curve! The problem perhaps isn't in the imaging system, but in the samples involved. I am hoping that someone could tell me why my approach involving LSF(x) had produced such a bizarre graph?
Many thanks