I am doing a radiation damage survey on a few different materials and will need the Detective Quantum Efficiency (DQE) when calculating the dose. I will be using the same detector for each sample and was wondering if it would be sufficient to acquire the DQE once, or even use a previously determined DQE, rather than recalculating it for each sample. My thinking is that since the detector is the same for each sample the DQE shouldn't fluctuate appreciably.

$\endgroup$
0
$\begingroup$

Short version: It's better to measure it in situ and shortly before (or better before and after) taking data.


Explanation

Quantum efficiency can be a function of various operating parameters, so you would need to know that the previous measurement was done using the same parameter you plan to use.

Various classes of detectors can experience changes in their actual quantum efficiency for various reasons: radiation damage; current erosion; mechanical, chemical, or thermal damage at a minimum.

It is also possible that the previous measurement intentionally or unintentionally folded some acceptance features into their reported QE.

For these reasons it is better to measure the QE

  • explicitly for the data taking you plan
  • in the configuration you will be using
  • under the operating parameters you will be using
  • in close temporal proximity to the data taking

But, as always, you have to balance the risk against and the costs. If you know that the device was previous testing in a configuration compatible with the one you intend under operating condition compatible with those you intend, and has been stored with due care to protect if from things that could damage it (which could include humidity, bright light, intense radiation fields, mechanical shocks, and others depending on the nature of the detector), and you don't have a lot of money/time then you might decide to chance it.

| cite | improve this answer | |
$\endgroup$
  • $\begingroup$ Any idea what kind of variance I would be looking at if I were to chance it? Or is that an unrealistic thing to guess at? $\endgroup$ – onb Mar 13 '18 at 20:15
  • $\begingroup$ That really depends on the type of detector, what might have gotten to it, and how severe the damage is. When things are going well these are usually small changes (a few percent of the previous QE, maybe up to 10%), but there isn't really a limit. $\endgroup$ – dmckee --- ex-moderator kitten Mar 13 '18 at 20:26